Update on the ASAIR AFS01 CO₂ flow sensor
I've been experimenting with the AFS01 (0–200 SCCM) using an ESP32 and ESPHome, integrated into Home Assistant. After obtaining the latest May 2026 documentation directly from Aosong and conducting some calibration tests, here's what we've established:
1. Technical specifications
- I²C address: 0x40
- Supply: 5V (requires a level shifter with ESP32)
- Accuracy: ±5% of reading; repeatability: ±1%
- Pressure resistance: 0.3 MPa (3 bar), although gauge vs absolute isn't specified
- Factory calibration gas: Air at 25°C and 1 atmosphere
2. Official SCCM conversion formula
The latest manufacturer documentation finally specifies the conversion:
Flow (SCCM) = (Raw Reading − 32000) / 140
The divisor of
140 applies specifically to the 200 SCCM model.
3. Experimental validation
I calibrated the sensor using a 20 mL syringe, repeating the exercise across multiple runs and integrating readings sampled every 500 ms.
My empirical conversion factors came out around
135–138, remarkably close to the manufacturer's official value of 140. This gives me reasonable confidence that the sensor measures air flow correctly.
4. CO₂ correction factor — the remaining challenge
I contacted Aosong directly. They confirmed that the AFS01 has
never been calibrated for CO₂, so they cannot provide a correction factor. They did confirm that its response is linear.
I subsequently researched correction factors from Bronkhorst, Aalborg, Brooks, Teledyne Hastings, Renesas and Honeywell.
Several manufacturers converge around
0.70–0.75, with
0.74 being especially common for CO₂ relative to air/nitrogen. However, scientific studies demonstrate that the factor depends on the sensor's thermal design, so this cannot be assumed to be accurate for the AFS01.
For now, my provisional CO₂ formula is:
CO₂ SCCM ≈ ((Raw − 32000) / 140) × 0.74
5. Current status
The ESP32/ESPHome setup is working, with stable readings visible in Home Assistant. Air calibration is complete, but the
0.74 CO₂ correction remains unverified for the AFS01.
The outstanding task is obtaining a reliable reference measurement of actual CO₂ flow to establish the sensor-specific correction factor.
If anyone has independently calibrated an AFS01 using pure CO₂, I'd be very interested in comparing results.